- From: jess <jess@sci.utah.edu>
- To: gorman.sydney@mayo.edu
- Cc: scirun-users@sci.utah.edu
- Subject: Re: [SCIRUN-USERS] Problems setting contact voltage on DBS lead
- Date: Tue, 25 Feb 2020 16:50:41 -0700
Hi Sydney,
There are a few reasons that there would be zeros in your solution. The most
common is that the voltages are not set properly. To calculate the potential
field from known electrodes (boundary conditions) the field need to be the
set voltages only at the desired locations and nan everywhere else. You can
test this using clipfieldbyfunction and using 'isfinite(DATA);’ in the
parser. Other things to check include the conductivities (cannot be 0 or too
high), and the error on the solvelinearsystem module (which should be 10e-6
or lower).
cheers,
Jess
>
On Feb 25, 2020, at 10:57 AM, gorman.sydney@mayo.edu wrote:
>
>
Hello!
>
>
I'm attempting to model the electric field produced by a DBS electrode
>
using a
>
voltage controlled model. I'm basing most of my current Sci Run environment
>
off of the example published in "Targeting Neuronal Fiber Tracts for Deep
>
Brain Stimulation Therapy Using Interactive, Patient-Specific Models.",
>
however, the results of the linear system seem to be zero regardless of the
>
voltage I have set. I was wondering if anyone had run into a similar
>
problem,
>
or if there is something obvious I am missing. I can send a picture of my
>
workstation if that would be helpful!
>
>
Thank you so much!
>
>
Sydney Hillan
>
>
>
________________________________________________
>
SCIRun users mailing list: scirun-users@sci.utah.edu
>
http://www.sci.utah.edu/software/scirun.html
>
To unsubscribe, email sympa@lists.sci.utah.edu with the "unsubscribe
>
scirun-users" in the message body.
Archive powered by MHonArc 2.6.18.